Which model Omega meter are you speaking of?
Omega PHH-7011
Which model Omega meter are you speaking of?
Hole (normally closed with a stopper of some sort) in the lid and a wine thief?
I've been thinking about the same issue... Would there be any reason that taking a sample from the spigot on the mash tun wouldn't work?
Yes.I'm going to assume the advice is still to store in the cap with some water.
Or RO water or DI water or 7 buffer,1) If my storage water dries up, what is the best way to rehydrate the bulb? 10-15 minutes in tap water?
You want fresh buffers each time you run a test but for a single test just put the electrode in 4 buffer, stir occasionally and record displayed pH.2) I'm using premixed buffers and I realize that for calibrating I will always want fresh buffers however for the stability test do I need fresh buffers for each testing?
You record every 5 minutes or so. What you are looking for is it stabilizing. It should stop stabilize after a few minutes. An erratic history is indicative on an unstable meter/electrode and one that is, therefore, innaccurate.3) How do I know when a reading is stabilized? When I put my meter into the 7 solution for calibration it started at 7.12 and gradually kept climbing down and never seemed to stop. I gave up when it was at 6.91.
Not a good sign. The sticky has a plot from a stable meter.Every time I thought it had stabilized it would drop another tenth. This was over the course of about 5 minutes. It never stayed on any given reading for longer than a minute though.
Yes, it is responsible for 'dilution error' but that shouldn't amount to a hill of beans. Turn the cup upside down and rap it on the counter. That should get 99.9% of the rinse water out.4) When rinsing with DI water, how important is it to get all the water out of the sample cup? I've been pretty OCD about it but I'm wondering if I need to be...is it the PH of any risidual water we're worried about affecting the buffer or the dilution of it?
Should be fine.As for the meter itself, I've been spraying it and as AJ mentioned just using a paper towel to wick at it without putting any pressure on the bulb.
Were the noise figure of your meter 0 and the electrode perfectly stable, yes. O/W no. A drift of 0.01 is perfectly acceptable as long as you stay near 3.99 for a half hour or hour or two.5) Once I calibrate 7 and 4 and save, shouldn't my reading be 4.00 if it's still sitting in the buffer? I was getting 3.99 (which I realize is within the accuracy range just thought it would show exact right after calibration).
Either.6) When I put the meter in the sample, should I move it around before locking the cap on or lock the cap on and then shake it around?
Don't believe this meter displays either. That is usually a feature of more expensive meters.7) I RTFM and didn't see anywhere that mentioned where we could find the slope/offset values on this meter...correct me if I'm wrong.
Yes, recal, or at least do a cal check, every time you use it. When the electrode wears out the meter will tell you it can't cal.If we don't know what it's using, how do I know that the meter is calibrated any given time I turn it on? Do I just plan on calibrating every day/time I use it?
Yes, because the pH of the buffers change with temperature. Try to hold buffer temp constant during a stability test.8) with ATC, do I need to be concerned with the temperature at all when recording the readings?
I have not found it necessaary even with electrodes I have used for years but that doesn't mean that you won't.9) Other than spraying with DI water right after use, is there any need for cleaning solution? I think I read in this thread it was unneeded but just wanted to verify.
DOH! Apologies for the stupid link trick (fixed).
This three solution kit is what I've been buying...
Cheers!
Don't know what GP's are. I'll point out, though, that your meter assumes linearity between buffers so that, for example, if you are checking mash pH there is no point in calibrating against the 10 buffer. Calibration data derived from 10 buffer are used only for measurements between 7 and 10 pH.My other meter (Hach PP+) does support 3 point and I do use the 10 solution for that meter just on GPs.