ajdelange: You said, "See the Sticky on pH meter calibration for tips." I don't see anything like that; is this posted somewhere else?
It's at the top of this page https://www.homebrewtalk.com/showthread.php?t=302256
ajdelange: You said, "See the Sticky on pH meter calibration for tips." I don't see anything like that; is this posted somewhere else?
I think you have to hold the arrow buttons down.
In my experience (and it took a while for me to figure this out) is that calibration takes a long time with this meter. I put it in the buffer, move it around a bit then let it sit until it recognizes the buffer. It takes a minute or two for each one.
sooooo... any more updates from the field, now that it's been in use for a while?
I have used mine on two batches now. I calibrate it each time as it gets off pretty fast but each calibration has been fast and agree with above post, gets a stable reading in about 20 seconds.
I currently use these from amazon to hold testing item and calibration fluid. https://www.amazon.com/gp/product/B00BJPGIPQ/?tag=skimlinks_replacement-20
I use 4 of them, one has RO water for rinsing and then two more for calibration fluid and one with the beer sample. The cylinders work great because you can set it in and walk away and it holds it up nicely and keeps it submerged.
Have you figured out how to toggle the readout up or down while in CAL mode? Instructions say to press up or down, but when I do that nothing happens.
Why this discrepancy between 6.86 and 7.0? Is the buffer solution actually 7.0 or 6.86? Others talk about 7.01. What's the point of 6.86 when we are trying to calibrate for 7.0? Or are buffer solutions slightly different, ranging between 6.86 and 7.01?
The pH of a buffer depends in its temperature. This is a basic fact of the chemistry. The meter's calibration procedure has to take the temperature of the buffer into account when it calibrates and has to, similarly, consider the temperature of the sample when it estimates the pH from the electrode voltage. Good modern meters have the table (or formulas) of buffer pH vs temperature built in and it should not, therefore, be necessary to make adjustment for temperature during calibration. I do not know anything about this particular meter.
Thank you. However, the difficulty is in understanding what pH this buffer actually is. In the charts, the basic pH rating is for 25 C, but they are 4.01, 6.86, and 9.18. The buffer solutions I purchased are for 4.0, 7.0 and 10.0, within 0.01 accuracy. If I assume these values are also at 25 C, they obviously differ significantly (especially the 10.0 buffer). The odd thing is that these buffer solutions come from the same manufacturer. In fact, in their table, under the 10.0 buffer, no temperature even comes close to the 10.0 value. The closest is at 0 C, which is 9.47.
I've ordered one of these so I'm quite interested in the conversation here.
Rhys, have you tried pressing and holding for 3-4 seconds? On my Inkbird controller, to reset parameters you have to hold the set button for more than 3 seconds for it to kick into reset mode, then have to hold it again for more than 3 seconds to save the changes.
Not sure that would be the problem here, just an idea.
An error of .01 on a cal check does not require correction.
Just a few questions about the buffer solutions.
1. What volume would be considered adequate enough for each calibration use? About 20 ml?
2. Can these smaller volumes be used more than once, if stored in a separate small bottle? Just wondering, because these buffer solutions are fairly expensive.
3. For the 8689 pH Meter, is it necessary to calibrate all 3 buffer solutions each time, especially if the upper range is not used? (i.e is there really any need to calibrate the 10.0 buffer?).
1. What volume would be considered adequate enough for each calibration use? About 20 ml?
Because a small volume is more easily contaminated relative to a large one I'd say no. Unless you buy buffers in the individual ketchup/mustard/mayonaise/soysauce packaging they aren't that expensive.2. Can these smaller volumes be used more than once, if stored in a separate small bottle? Just wondering, because these buffer solutions are fairly expensive.
The meter assumes the pH is a linear function of the voltage produced by the electrode. In calibrating with a pair of buffers it determines the slope and offset of the calibration 'curve' between the pH's of the two buffers. Those values are not applicable to measurements taken outside the range spanned by the two buffers. So if your measurements are restricted to the range 4 - 7 pH then a two point cal with 4 and 7 buffers will suffice. You can accurately measure with a 4 - 7 cal a little outside the 4 - 7 range, say up to 8, but for best results you should use a 7 - 10 calibration for anything with 7 < pH < 10.3. For the 8689 pH Meter, is it necessary to calibrate all 3 buffer solutions each time, especially if the upper range is not used? (i.e is there really any need to calibrate the 10.0 buffer?).
Thank you, rhys33 and ajdelange! That is very helpful.