And then I saw the rest.
Yep so i am firmly in alignment now that the calibration routine on this meter is trash, and there's no sense attempting to use it any further. With that in mind i guess my only option is to manually correct the readings after doing my own calibrations...
1. Delete calibration from meter - it will never say 'CAL' on the front.
2. On brew day i will remove the cover from the probe, rinse it with tap water to get the salt off, and then give it a final rinse in DI water .
3. Pour a pH7 buffer and soak/swirl meter for 10 minutes. Record reported value.
4. Perform cleaning procedure (rinse/shake/dry)
5. Pour a pH4 buffer and soak/swirl meter for 10 minutes. Record reported value.
I'll then calculate a calibration curve and apply it to any of my measurements.
All of this still assumes my probe is good.
You can see in the last 2 pictures the calibration had been deleted again (it doesn't say CAL on the display) and the probe was reading nearly 4.1 when it should have been reading 4.00.
I'll perform the drift/stabilization test as you describe.