In short, yes you can, but here is why it won't be helpful, pasted from an email I received in response to a similar inquiry by the folks who make the coffeelab refractometers (or rather commissioned their construction)...
Its easy to empathize. The reason for the cost is that the refractometer is a custom product that resulted from a 2-yr and ongoing R&D collaboration between VST and our OEM provider, whereby the basic instrument has been re-designed and significantly improved in order to have the resolution, accuracy and precision necessary to be useful for coffee measurements. Typical off-the-shelf refractometers have accuracies that are limited to approximately ±0.15% or worse (ref index of ±0.0003 vs 0.00003), due mostly to the limited sensor array being 128 pixels, resulting in a refractive index resolution of 0.0001 and temperature resolution to 0.1 Deg C. This is not sufficient for coffee measurements. In addition, typical instruments can only warrant an accuracy of temperature to± 1.0 Deg C However, an error of ±1 Deg C is affects TDS ± 0.05-0.06%, by itself, and the accuracy errors are additive. The VST LAB Coffee III instrument resolves refractive index to better than 0.00001 using a 1024 element sensor and temperature to 0.01 Deg C.
The problem is that if the measurement accuracy is off by ±0.15% or more, then the extraction yield, the crucial parameter you need, can be off by > ±2%, making the measurement useless. In fact, the measurement error could cause you to adjust the brewing process in the wrong direction, actually making the the error in brewing protocol worse!
See the attached chart to see what that means. ±0.15% could be anywhere within the Red circle on the attached Universal Brewing Control Chart for a typical drip batch. The VST instrument is warranted for accuracy to be better than ±0.03%, but is contractually delivered to VST at ±0.02% or better (as shown in the blue circle on the attached Universal Brewing Control Chart). Typically, they measure ±0.01% TDS within the coffee range.
This kind of accuracy and precision are typically found only on laboratory bench top instruments using peltier trays (that maintain Temp of the sample and optics to ±0.02 Deg C) and cost > $10,000. That VST can deliver similar performance for under $1000 is a actually remarkable, and this is why it has become a world-wide standard shipping to some 75 countries and is used by professionals and amateurs alike, as well as for R&D and Equipment design by the worlds leading cafes, roasters and equipment developers.
Yes, we do receive units back that are lightly used at World Coffee Trade Events, competitions, samples to reviewers, and loans made to educational institutions. When these come up, they are usually already spoken for, but we can notify you should one become available.
Anyways... that's the story....
Of course if you like what you taste in the cup all is good.