I have started use a pH meter when brewing, and I had an issue brewing yesterday. Some background, I use a Hanna 98128 meter that is stored properly and have no problem calibrating. I also have Bru n' Water to use and a report from Wards.
When I started yesterday and had the meter calibrated ( 4&7 buffer ) I had the bright? idea that I wanted to check the pH of my water BEFORE I added the lactic acid that was needed. So I get a sample of water to test and the meter doesn't stabilize as I watch the display go a full point higher then my pH of 7.8. Took the meter out, cleaned it, tried again and same result as before. So I add the lactic acid and pray because now I'm NOT sure what the pH really is, even though I have the water report. So I'm thinking there is something wrong with the meter, and that maybe my probe is bad. After some searching, I read that when I am calibrating the meter with 4 & 7 buffers it is basically bracketing the meter's calibration for the solution intended to be tested.
So my questions: If I read that correctly does that mean if I want to check the pH before additions I would need use 7 & 10 buffers? And then re-calibrate with 4 & 7 to check mash pH? Do any of you check your water before or do you just take the number from the water report and roll with that? Or is just possible that my probe is bad and I need a new one?
When I started yesterday and had the meter calibrated ( 4&7 buffer ) I had the bright? idea that I wanted to check the pH of my water BEFORE I added the lactic acid that was needed. So I get a sample of water to test and the meter doesn't stabilize as I watch the display go a full point higher then my pH of 7.8. Took the meter out, cleaned it, tried again and same result as before. So I add the lactic acid and pray because now I'm NOT sure what the pH really is, even though I have the water report. So I'm thinking there is something wrong with the meter, and that maybe my probe is bad. After some searching, I read that when I am calibrating the meter with 4 & 7 buffers it is basically bracketing the meter's calibration for the solution intended to be tested.
So my questions: If I read that correctly does that mean if I want to check the pH before additions I would need use 7 & 10 buffers? And then re-calibrate with 4 & 7 to check mash pH? Do any of you check your water before or do you just take the number from the water report and roll with that? Or is just possible that my probe is bad and I need a new one?