Well-Known Member
I used my Hana pHep 5 (98128) last week by recalibrating with pH 4.01 & 7.01 buffers before each measurement, taking the measurement, and rechecking the calibration in the buffers for drift. In <10 minutes the meter read 4.00 & 7.05. That's ok drift for me and measurements mid-span are hardly affected. I have a spreadsheet tool for adjusting the actual reading to the post calibration check, but frequent calibration makes corrections very small.
Budget $40 for calibration buffers (4.01 & 7.01), electrode storage solution & electrode cleaning solution purchased with the meter. My meter is about 10 months old, only read room temperature samples, always kept vertical in cap full of storage solution, monthly cleaning solution and has proven reliable. Hana replacement electrodes are about $40 when the time comes.
Budget $40 for calibration buffers (4.01 & 7.01), electrode storage solution & electrode cleaning solution purchased with the meter. My meter is about 10 months old, only read room temperature samples, always kept vertical in cap full of storage solution, monthly cleaning solution and has proven reliable. Hana replacement electrodes are about $40 when the time comes.