The buffers are at the labeled pH values only at 25°C. At other temperatures their pH's are slightly different. There should be a chart on the buffer package that tells you what the pH is at various temperatures. If you calibrate and then measure at 25° C you should get pretty darn close to what the buffer was labeled as long as your reading is between (or equal to) the buffers used for the calibration. IOW calibrating with 4 and 7 buffer should get you what the buffer label says if you go back and measure 4 and 7 buffer. But you can't expect that with 10 buffer and a 4 &* 7 cal,