Sudz
Well-Known Member
I just received a new refractometer which I calibrated in distilled water with my old standby hydro. I also cal'd a second hydro a friend gave me and was surprised to see it was six points higher than my old one.
I mixed up some sugar water and adjusted it's temp to 60*F. I started with 1.090 and measured each device and wrote down the data as things progressed. I changed the concentration by randomly adding water and taking another set of data points. I did this repeatedly to gather a progression of data across a useful range.
I found the two hydros tracked the same from 1.010 to the 1.090 although the newer one held the 6 point margin throughout the range. The refrac tracked within 1 point of the old hydro after cal at 1.000. However, once readings moved above about 1.040 the refrac started to drop points. By 1.090 the refrac was reading 1.081.
This was a bit surprising to me. I hadn't expected to need to use a calibration curve when using the refrac. Bummer.
Have any of you observed this with your refractometers or do I just have a problem device?
I mixed up some sugar water and adjusted it's temp to 60*F. I started with 1.090 and measured each device and wrote down the data as things progressed. I changed the concentration by randomly adding water and taking another set of data points. I did this repeatedly to gather a progression of data across a useful range.
I found the two hydros tracked the same from 1.010 to the 1.090 although the newer one held the 6 point margin throughout the range. The refrac tracked within 1 point of the old hydro after cal at 1.000. However, once readings moved above about 1.040 the refrac started to drop points. By 1.090 the refrac was reading 1.081.
This was a bit surprising to me. I hadn't expected to need to use a calibration curve when using the refrac. Bummer.
Have any of you observed this with your refractometers or do I just have a problem device?