Bought a Hanna pHep5 98128 and have used it on two beers trying to get my mash pH right. I'm not entirely happy with it because right after I calibrate it, if I put it back in the 7.01 solution it can be off by as much as .1 (specs say that it's accurate to .05) then drift over the next 5 minutes.

I've never used a pH meter so I'm hoping someone can take a look at this and tell me if what I'm seeing is normal or if there's something wrong with this one. I will say that when I measure actual mash pH, it does not seem to drift, but because of what I see when measuring the calibration solution, I'm wondering if those readings are accurate.

The measurements below were taken after calibrating the meter in 7.01 and 4.01 solutions. After each test the probe was rinsed with water & as much liquid as possible blotted off.




Edit:Not sure if everyone else is seeing the red x instead of the screenshot, but here's what's shown when measuring 7.01 solution after calibrating:

Time pH

Initial 7.11
5s 7.13
10s 7.11
15s 7.10
30s 7.09
45s 7.08
1m 7.07
1.5m 7.05
2m 7.04
2.5m 7.04
3m 7.03
3.5m 7.02
4m 7.02
4.5m 7.01
5m 7.01
10m 7

__________________

Student7 is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 03:36 AM   #2
ajdelange
HBT_LIFETIMESUPPORTER.png
Feedback Score: 0 reviews
Recipes 
 
Join Date: Aug 2010
Location: McLean/Ogden, Virginia/Quebec
Posts: 6,372
Liked 662 Times on 549 Posts
Likes Given: 19

Default

pH meters take a while to respond. Note that what you have here is an indication which is initially off by quite a bit but which gradually migrates more or less monotonically to the correct answer. The rate at which an electrode comes to equilibrium is one measure of the quality of the electrode. One would expect an electrode which costs several hundred dollars by itself to come to equilibrium faster than 10 minutes i.e within a minute or so but it is usually always necessary to wait at least a minute. Ten minutes does seem like a long time but note that you are within 0.02 pH of the correct answer within 3.5 minutes or so. More expensive instruments monitor the rate of reading change over time and typically signal with a beep that the rate of change of pH per unit time has slowed to a value which can usually be set by the operator thus signalling that equilibrium has been approached. In the lab I have the meter connected to a computer which plots the readings over time so that I can eyeball the approach to equilibrium.


A slowing of the electrode response as the electrode ages is normal and a sign (along with greater offset and reduced slope) that it is aging. It can also signal that the electrode is dirty. Neither should be the case with a new instrument.

So yes, what you are seeing is normal in the sense that electrode responses drift finally settling out close to the right answer but it does seem a bit odd that it is taking so long. OTOH you are within spec at a minute and a half.

__________________
ajdelange is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 03:53 AM   #3
sudsmcgee
HBT_SUPPORTER.png
Feedback Score: 0 reviews
 
sudsmcgee's Avatar
Recipes 
 
Join Date: May 2010
Location: Western Chicago Burbs, IL
Posts: 1,519
Liked 49 Times on 41 Posts
Likes Given: 52

Default

How do clean an electrode if it's dirty?

__________________

"If you're here for the premature ejaculators meeting, you've come too soon."

sudsmcgee is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 04:01 AM   #4
Student7
Feedback Score: 0 reviews
Recipes 
 
Join Date: Sep 2011
Location: Knoxville, TN
Posts: 22
Default

Thanks for the response AJ. There were a couple other trials in the screenshot that didn't show up and I'm wondering if the 4.01 test still fits with this meter being ok but slow. In this case it seems to start close to the proper reading then move away from it.

Also checked vineagar and the differnce in readings over time is much larger

4.01 solution
Time PH
Initial 4.03
5s 4.02
10s 4.01
15s 4.01
30s 4.00
45s 3.99
1m 3.98
1.5m 3.98
2m 3.96
2.5m 3.96
3m 3.95
3.5m 3.95
4m 3.94
4.5m 3.93
5m 3.92

Distilled white vineagar
Initial 2.5
5s 2.46
10s 2.45
15s 2.45
30s 2.42
45s 2.4
1m 2.37
1.5m 2.34
2m 2.31
2.5m 2.29
3m 2.27
3.5m 2.26
4m 2.25
4.5m 2.24
5m 2.23

__________________
Student7 is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 04:04 AM   #5
Student7
Feedback Score: 0 reviews
Recipes 
 
Join Date: Sep 2011
Location: Knoxville, TN
Posts: 22
Default

Quote:
Originally Posted by sudsmcgee View Post
How do clean an electrode if it's dirty?
I used the manufacturers cleaning solution and soaked the electrode in it for an hour then rinsed it. Is there anything else should be done to clean it?
__________________
Student7 is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 05:10 AM   #6
ajdelange
HBT_LIFETIMESUPPORTER.png
Feedback Score: 0 reviews
Recipes 
 
Join Date: Aug 2010
Location: McLean/Ogden, Virginia/Quebec
Posts: 6,372
Liked 662 Times on 549 Posts
Likes Given: 19

Default

WRT to cleaning electrodes: the best method is to do what the manufacturer recommends. That said, in brewing deposition of protein on the electrode can be a problem so a soak in an protein lysing enzyme is often a good idea. There are products made especially for cleaning brewing electrodes but things like Zymit should work.

WRT drift: I have tentatively concluded that one mark of the less expensive meters is that they drift more than the more expensive ones. Looking at the data for 4 buffer and vinegar note that after a minute or so the readings may be subject to variation but that they are reasonably tightly grouped. The standard deviations are less than the specified accuracy of the meter. Keep in mind that the accuracy of a sample reading depends on threereadings: one for each of the buffers and one for the sample. If you know that your electrode response is slow you should wait a good long time in each buffer before accepting the reading as part of the calibration process.

__________________
ajdelange is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 06:43 AM   #7
samc
Feedback Score: 1 reviews
Recipes 
 
Join Date: Aug 2008
Location: Portland OR
Posts: 5,420
Liked 55 Times on 52 Posts
Likes Given: 29

Default

May be of some use - I bought a cheap PH meter off EBAY with very poor performance. Put in some better quality batteries and the performance was vastly improved. Might be worth a shot.

__________________
samc is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 02:13 PM   #8
broadbill
Feedback Score: 0 reviews
Recipes 
 
Join Date: Aug 2007
Location: Southern Maine
Posts: 3,408
Liked 322 Times on 226 Posts
Likes Given: 224

Default

Regardless of the fact that the meter is not quite as accurate as the specs would indicate...the variation it is showing is still small enough that is it OK for brewing purposes, even good enough for research laboratory standards.

I wouldn't worry that much....it is probably due to electrode quality as ajdelange stated. Is the 0.05 spec listed by the meter or by the electrode?

__________________
broadbill is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 03:00 PM   #9
ajdelange
HBT_LIFETIMESUPPORTER.png
Feedback Score: 0 reviews
Recipes 
 
Join Date: Aug 2010
Location: McLean/Ogden, Virginia/Quebec
Posts: 6,372
Liked 662 Times on 549 Posts
Likes Given: 19

Default

pH meter accuracy is a pretty involved subject as it depends on several factors
1. Drift in electrode offset
2. Noise in electrode response
3. Drift in electronics
4. Noise in electronics (including quantizing noise in the A/D)
5. Accuracy of the buffers used for the calibration (i.e. are they 4.01 pH ± 0.02 or ± 0.05?).
6. Accuracy, noise and bias in the temperature measurement. Either ATC is used or the buffers and sample must be held at the same temperature. In either case there is error associated with the temperature measurement.
7. The pH of the sample relative to the pH of the buffers. Best accuracy is obtained when the pH of the sample is half way between the pH of the buffers. It falls off quickly when the pH is outside that span.

In a good laboratory instrument the design is such that the buffers' accuracy is the limiting factor (at least for the usual NIST ± 0.02 pH buffers) so that a reading made mid span is usually accurate ± 0.01 pH.

Note that the mean of the pH 4 buffer readings in this case is 3.9755 and the standard deviation 0.034. Thus the standard error in the mean is 0.009. Thus the pH estimate based on the simple average of the measurements is 3.9755 ± 0.001 i.e. between 3.974 and 3.976 i.e. within 0.05 of the actual buffer pH so the meter is within spec. OTOH there is a definite downward drift over time. One would ordinarily discard any readings taken in less than a minute. If you discard those here you would get an average reading of 3.956 ± 0.007. Meets the 0.05 accuracy at the high limit and just misses it at the lower. Had the 4 buffer calibration been done after waiting 5 minutes I'm guessing the 4 buffer reading would be within the accuracy spec.

As you can see it's a bit of a numbers game. Experience teaches users how to interpret pH meter readings properly. Until then it is quite possible to be led down the garden path by one.

__________________
ajdelange is offline
 
Reply With Quote Quick reply to this message
Old 10-24-2011, 09:50 PM   #10
Student7
Feedback Score: 0 reviews
Recipes 
 
Join Date: Sep 2011
Location: Knoxville, TN
Posts: 22
Default

Quote:
Originally Posted by broadbill View Post
Regardless of the fact that the meter is not quite as accurate as the specs would indicate...the variation it is showing is still small enough that is it OK for brewing purposes, even good enough for research laboratory standards.

I wouldn't worry that much....it is probably due to electrode quality as ajdelange stated. Is the 0.05 spec listed by the meter or by the electrode?
Thanks for the reply and the reassurance, I was kind of considering selling it and getting the Milwuakee MW102, but I think I'll keep it for now and see how it goes. Still, it would be nice to know when it says my mash pH is 5.3, is that really 5.35 or 5.25 or somewhere in the middle?

The .05 accuracy is from the meter's spec sheet. I can't find anything on the manufacturers site showing the specs of the probe itself.

Quote:
Originally Posted by ajdelange View Post
Keep in mind that the accuracy of a sample reading depends on threereadings: one for each of the buffers and one for the sample. If you know that your electrode response is slow you should wait a good long time in each buffer before accepting the reading as part of the calibration process.
Unfortunately this meter autocalibrates and does not let you choose when the reading is accepted. Your explanation does help me understand (I think) why I'm seeing different behavior measuring the 7 buffer vs the 4. Likely when it autocalibrated on the 7, it was close to equilibrium but with the 4 it was not. When I later measure the 4 solution as the meter is approaching equilibrium, the reading moves past the calibration point and keeps going because the calibration point was accepted too soon.

Quote:
Originally Posted by ajdelange View Post
As you can see it's a bit of a numbers game. Experience teaches users how to interpret pH meter readings properly. Until then it is quite possible to be led down the garden path by one.
That's what I'm trying to avoid and I appreciate your help. At this point I think the way I'll look at it when I measure mash pH is that the reading after 1-2 minutes is likely within .02-.03 of the actual pH and more likely to be slightly higher than actual pH (based on the 4.0 calibration point being accepted too soon).

I guess another thing we have going for us is that typical mash pH fits in nicely near the halfway point between calibration points.
__________________
Student7 is offline
 
Reply With Quote Quick reply to this message
Reply



Quick Reply
Message:
Options
Thread Tools


Similar Threads
Thread Thread Starter Forum Replies Last Post
Recommend a pH meter? gbx Brew Science 10 10-24-2011 11:40 PM
Using a pH meter nilo Brew Science 9 06-07-2011 01:14 AM
Looking for PH Meter Suggestions RonRock Brew Science 43 04-20-2011 01:04 PM
Proper use of a pH meter EvilBrewer Brew Science 18 12-21-2010 07:01 PM



Newest Threads

LATEST SPONSOR DEALS